HP/Agilent 5890 Warning: Oven Shut off
Discussions about GC and other "gas phase" separation techniques.
26 posts
Page 2 of 2
Sounds like you're doing some good troubleshooting. Good luck getting to the bottom of it!
-
- Posts: 159
- Joined: Thu Jan 05, 2012 5:27 am
Mea culpa. I had thought there was an actual switch there, but I got into mine last night to find out I was quite wrong. My apologies.
The best I could do was reproduce the error by leaving the door ajar, resulting in an inability to ramp the temperature up fast enough, so- nothing new there.
-
- Posts: 373
- Joined: Mon Sep 13, 2010 8:28 am
There is diagnostic/tuning mode for heaters (entered pressings [Stop] during boot) or using [.] commands (can not remember exactly).
When mainboard lost PID constants for heater - you can get such problem.
Try to re-calibrate oven sensor and heater.
When mainboard lost PID constants for heater - you can get such problem.
Try to re-calibrate oven sensor and heater.
-
- Posts: 159
- Joined: Thu Jan 05, 2012 5:27 am
And for those playing the game at home, that's on page SVC 5-29:Try to re-calibrate oven sensor and heater.
https://www.ursinus.edu/live/files/140- ... -manualpdf
-
- Posts: 11
- Joined: Mon Dec 07, 2015 6:10 pm
Thanks for the further replies.
the oven and other temperature sensors measure room temperature reasonably.
I can only do a one point (re)calibration at room temperatures, since I can't heat up the oven.
I adjusted delta T by 0.1 K.
It didn't solve the problem.
the oven and other temperature sensors measure room temperature reasonably.
I can only do a one point (re)calibration at room temperatures, since I can't heat up the oven.
I adjusted delta T by 0.1 K.
It didn't solve the problem.
-
- Posts: 373
- Joined: Mon Sep 13, 2010 8:28 am
Does the oven thermal sensor show adequate value when you heat it with the solder (~350C)?Thanks for the further replies.
the oven and other temperature sensors measure room temperature reasonably.
I can only do a one point (re)calibration at room temperatures, since I can't heat up the oven.
I adjusted delta T by 0.1 K.
It didn't solve the problem.
Does the value increase monotonically when you heat the sensor or jumps?
-
- Posts: 11
- Joined: Mon Dec 07, 2015 6:10 pm
Currently the main PCB is removed from the GC so I can't observe the readings. I will respond on that as soon I replace it.
*** Update
We put the 5890 Series ii board into a working 5890 series i and could reproduce the error:
Oven not heating
Fan not running.
While playing around with the working 5890 series i and it's original main PCB, we found out that the contactor on the Power PCB in the back of the GC (picture on post from DEC 7th) gives a physical response to entering a temperature on the keypad. It clicks and creates a contact, you can both see and hear it. Again, as posted earlier, this contactor is main PCB controlled.
Right now, we are hunting for a replacement board, since this test again hints towards a main PCB issue.
I will give an update.
*** Update
We put the 5890 Series ii board into a working 5890 series i and could reproduce the error:
Oven not heating
Fan not running.
While playing around with the working 5890 series i and it's original main PCB, we found out that the contactor on the Power PCB in the back of the GC (picture on post from DEC 7th) gives a physical response to entering a temperature on the keypad. It clicks and creates a contact, you can both see and hear it. Again, as posted earlier, this contactor is main PCB controlled.
Right now, we are hunting for a replacement board, since this test again hints towards a main PCB issue.
I will give an update.
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
I've got a fair amount of used 5890 parts from working units which were disposed of because they were no longer needed. Reasonable too, if shipping to within USA; but I can only do PayPal, money order or check, no credit card.Right now, we are hunting for a replacement board, since this test again hints towards a main PCB issue.
Post the part number(s) of what you are looking for, maybe even a photo too, along with your E-mail address. I'll post if I have that part, and I'll send you an E-mail asking you to make an offer.
-
- Posts: 11
- Joined: Mon Dec 07, 2015 6:10 pm
***update
swapped in replacement mainboard - system back running, no FAULTS, all heated zones get powered up.
visually, the faulty mainboard looks totally fine.
thanks for the offer KM-USA, unfortunately I read it after hunting on ebay. We have 3 5890s running, so who knows, what comes next.
Thanks to everyone for the steady support. I'll probably give another performance update after a couple of weeks.
swapped in replacement mainboard - system back running, no FAULTS, all heated zones get powered up.
visually, the faulty mainboard looks totally fine.
thanks for the offer KM-USA, unfortunately I read it after hunting on ebay. We have 3 5890s running, so who knows, what comes next.
Thanks to everyone for the steady support. I'll probably give another performance update after a couple of weeks.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Three operating 5890 units? Maybe buy his entire stash of parts.We have 3 5890s running, so who knows, what comes next.
-
- Posts: 102
- Joined: Wed Feb 22, 2012 7:24 pm
Glad to hear you finally got to the bottom of it. Thanks for the update!
26 posts
Page 2 of 2
Who is online
In total there are 68 users online :: 2 registered, 0 hidden and 66 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: BB65, Google [Bot] and 66 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: BB65, Google [Bot] and 66 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science