I'm using an old Varian 3800 GC with a Saturn 2000MS for pesticides analysis. We have quite a few of these within our laboratory but with I'm currently expericing a repeated issue with the CP-8400 autosampler (which isn't the best designed autosampler in the world!) on one of my instruments. This particular 8400 is repeatedly crunching the syringe needle several vials into the run and I'm at a loss as to what's causing it.
Some details of what I've tried so far:
- Each time I replace the syringe, I recalibrate the vial 0 and injector port positions. Alignment seems to be correct and it successfully injects several samples (between 8 and 10) before crunching.
- I replace the septum that sits within the injector port each run, so I doubt that it's the septum hardening with age!
- I've moved the samples onto another GC with an 8400 autosampler and that has injected without any problems so I don't think it's the vial cap septa.
If anyone has any ideas, please help!
All the best,
Aviator
I am in quite the same boat, but have had ZERO problems with carousel injector calibration & programming. The poltergeist that has irreversibly (programmed into itself is the injector position location). Used to be that I could use the 8400 to use either the front or middle
positions as long as I selected "Both" on the AS menu on the 3800. Nowadays, it only recognizes injector "1" front position as the correct position (#2) that I wish to use. This occurred after a simple column change. {the logic "magically reset itself last week after the 8400 was accidentally dropped on to my Dell CRT screen}..but only lasted for the remainder of that day,
No former Varian service tech. has been able to solve this weird problem, but being that I am only interested in the middle inj. pos., the calibration proceeds nicely, EXCEPT for the fact that I ruined ~10 syringes last week, even at 5 of a max 30cm/sec inj speeed at only 75% needle depth.
Give me a tech exorcist or the fat psychic from "Poltergeist"!
I am currently about to try pre-cored, softer septa from Restek & perhaps a slightly larger (& stronger 23 gauge needle, even if I have to re-machine the bottom brass/gold bottom alignment pin on the bottom of the sled. The Power of Christ compels me to scream at the engineers of this piss-poor machine.