Normal background levels on Waters Xevo TQ-S?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
We have a Waters Acquity UPLC/Xevo TQ-S (new instrument to me) and we are seeing very a very high background noise injecting pure solvent/mobile phase (5 to 6e5 TIC). We have rinsed the UPLC using the recommended solvents, and doing direct infusion via the IntelliStart-fluidics gives the same result. I have no real idea of what has been injected into the system previously. The cone has been cleaned. Any tips on what I should look at next? We're using proper solvents. :)
Hi
Let us know what are you doing? If you're scanning in 100-500 amu range, this can be normal. If you're doing MRM on a specific Ions, this is not good.
Thanks for your reply,
Currently I'm doing a regular UPLC -> TIC LC-MS scan to determine the ID/retention times for different analytes. Latter on, it will be purely MRM work. Yesterday I managed somehow to get the system "clean", with a pretty OK baseline using 100% CH3CN, but today it looks like the noise is back up again (waiting to see if it stabilizes now). So the culprit might be the H2O we're using (MilliQ). I understand that it's a sensitive instrument and since the guy who worked with it previously left I can't really compare the background levels with anything. Sometimes it goes up to 1-2e8 which really shouldn't be normal, right? Even when doing full scans. Talked to Waters as well, and they say the system might need a good cleaning.
Do a low range scan ( 50-150) and see which ions are dominant. then you can guess what is main source of problem. then, do a high range scan ( 500-150) and see if there is a big ion there or not.
Hi,

We never analyze in full scan mode on our TQ-S, only in MRM. However, before we start an acquisition, we check noise on the tune page at 95% H2O (+FA 0.05%). The scan time is set at 0.4 s and the range is 100-1000 Da. The biggest peaks are in the 10e7 range, so I am not that surprised about your background noise.
5 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry