Precision using ECD
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
-
- CE Instruments
What is your signal to noise ratio for the peak ?
If I calculate correctly 1ppm solution injected On-column would be equivalent to 100pg on column
From the specifications on a Thermo GC8000 from 1992 we had a signal to noise ratio of over 100:1 for 2pg on column for Chlorinated pesticides. If this is the correct quantity on-column it sounds as though either you have appalling reproducability or very poor sensitivity or a combination of both.
You do not state which chlorinated pesticide, not all have such a good response on ECD. To check the system I suggest you look to the manual find out the quoted performance specifications for the Clarius and then run the test standard. 8% rsd may be acceptable if you are running at minimum detction limits but 1ppm is not MDL for most Chlorinated pesticides.
For reference MDL Lindane 0.1pg on column ECD40/ECD400 in 1992
If I calculate correctly 1ppm solution injected On-column would be equivalent to 100pg on column
From the specifications on a Thermo GC8000 from 1992 we had a signal to noise ratio of over 100:1 for 2pg on column for Chlorinated pesticides. If this is the correct quantity on-column it sounds as though either you have appalling reproducability or very poor sensitivity or a combination of both.
You do not state which chlorinated pesticide, not all have such a good response on ECD. To check the system I suggest you look to the manual find out the quoted performance specifications for the Clarius and then run the test standard. 8% rsd may be acceptable if you are running at minimum detction limits but 1ppm is not MDL for most Chlorinated pesticides.
For reference MDL Lindane 0.1pg on column ECD40/ECD400 in 1992
2 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 17 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science