I bought a Trace Ultra GC with FID and TCD (used without software), and I have got problems with data collection with chromeleon 7.2 ( I have got one with trace 1300)
When I am using background monitor everything is ok but when I started analysis signal from both detectors is jumping from 10mV to 3000mV. On GC screen everything looks fine but in chromeleon ..;/
I have tried three computers.
I don't know if it is software problem? or hardware?
Thanks for your advice!
Hi Marek2
I've several questions to you:
- Are you able to control the instrument from the panel correctly?
- What do you mean with "Background Monitor"? Manual acquisition?
- What about the audit trails? Any errors written in?
- Did you configure the instrument in the Chromeleon server configuration by your own and are there maybe any errors in?
Wishing you success and best regards,
Stefan
Stefan Brand
Thermo Fisher Scientific