SPME headspace extraction with PDMS
Noticed in a recent project that these compounds were gone.
Have tried:
Removing a loop at the injector port
Removing the column end at the MSD in case any active sites had formed there
Switching to DVB/CAR/PDMS fiber for better extraction of these compounds
Changing liner (use 0.75 mm Supelco)
I had been cleaning and reusing liners but this last one was an old but never used liner I had in a drawer
I’ve been testing with a 100 ppb aqueous solution of these compounds and running headspace. I can’t even see them with SIM. I have ethyl butyrate in the mix and that always comes through fine. Instrument is a 7890A/5975C
Any suggestions would be appreciated