lmh » Mon Nov 09, 2015 3:15 pm
There are several things that can go wrong with a LOD based on SE of calibration curve. If your curve was taken over an inappropriate range of concentrations (for example far higher than the LOD) then the error at the LOD may be very badly over-estimated, in which case the LOD is wrong (too high). As James said, if the response curve over the range you used wasn't linear, the situation is awful, because much of the "error" isn't really random error at all, but is a systematic deviation from a straight line.
The idea of basing LOD on error of the calibration curve is that this error is your measure of how likely it is that the peak would, through statistical random fluctuations, disappear below the threshold of visibility. If the error isn't statistical random fluctuations, the whole argument goes wrong.
I am deeply suspicious of statistical calculations that tell me something that my eyes and common sense tells me is wrong. Nevertheless I don't think you can ignore your LOD calculation just because it doesn't make sense. Really, you are quite right to treat this as a trigger that something is going on, and needs to be understood.
There are, for example, situations where the LOD is telling you the truth: that although you can see the peak in a few samples at low concentration, its area is actually so hugely random that there's insufficient guarantee you'd see it in the next sample.
To take a very extreme situation, imagine an autosampler with a very badly-washed needle and a "sticky" analyte. It's likely that there will be a nice peak clearly present in all injections, but its area will be very much a random number, depending as much on the previous sample as the current one! In this situation S/N tells you that the peak is detectable (even in the blank if you ran one! LOD = 0.00, and a lovely graphic to prove it!), but a LOD based on the SE of the calibration curve is going to be very high, because the calibration curve has vast errors.
Another situation: imagine a mass spec doing a method that causes rapid build-up of source contamination, leading to loss of efficiency. The peak areas of three injections of the lowest level standard are 100, 60 and 30. Even in "30" the peak is very obvious with a nice S/N ratio, but quite rightly, the SE-of-curve method knows that the SE is very big and the area of the next injection could easily be 5, or even -5; with this trend, you can't guarantee to detect the next sample.
I'm pretty sure you'd spot these extremes though!