This will avoid: 1) measuring the pH in organic (which is not accurate) and 2)contaminating the eluent with the pH probe
Extremely small deviations in pH will still be present, but given that a good method chooses a buffer pH far from the analyte pKam this should not be a problem. At the same time you will be gaining extra reproducibility from constant organic/water mix