LC/MS/MS (ESI)-When is it appropriate to quantify Na adduct?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

17 posts Page 2 of 2
Thanks for your input, Klaus. I'm returning to this project soon, and will be doing some MS optimization again as I'm trying to add some additional compounds to the method. I'll keep an eye out for that possibility.
I've been running LCMS methods for Paclitaxel for almost a decade. About 90% of the time as a Na adduct, on Sciex 4000 and 5000 instruments. For some odd reason I've never been able to figure out, I have one instrument with much better response as M+H

I've never had issues with the Na adduct using a labeled internal standard. However, I did have one project where I used an analog internal standard that did not ionize as a Na adduct. It had pretty bad injection reproducibility, but that was only one study out of almost 100
17 posts Page 2 of 2

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry