Method Validation Question

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I am developing a test method for a pharmaceutical analyte that will calculate its purity by simply monitoring the % area of the primary peak within the chromatogram. Can anyone offer suggestions as to which of the usual validation parameters can/should be employed in validating a % area test method?
First and foremost: specificity. In the case you are proposing, you should establish that the response factor for you analyte is comparable (ideally, identical) to the response factors of likely impurities/contaminants. For system suitability, I would look at resolution from the closest impurity/contaminant and tailing factor. The rest is pretty much standard: linearity, accuracy (recovery), precision, etc.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 100 users online :: 0 registered, 0 hidden and 100 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 100 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry