Using water matrix for construction of calibration curves

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

19 posts Page 2 of 2
Thank you Peter.

Can I know why is it the case? Is it because the recovery of the analyte might be incomplete or enhanced by the matrix?
It is because the fraction of the analyte that partitions into the headspace, is taken as part of the headspace sample, and ends up at the detector is the same as the fraction of the standard that you add as the standard additions. The "recoveries" for native analyte and added standard are identical - so they cancel out of any equation in which you might want to include them.

One of the great strengths of the method of standard additions is that the matrix for the sample and the standard are identical, because you are creating standards by spiking portions of the sample. Under these conditions there is no need for the corrections for recovery that are required to correct for matrix effects when standards are not made up in sample matrix.

Peter
Peter Apps
Also if doing Method of Standard Addition Calibration, then you are calibrating out the unknown amount in the matrix to determine a response factor for a known amount of analyte.

If you get a response of 25 for the matrix alone at the detector, then you add 10ppm standard and you now get a response of 50, then it would calculate out that the unknown concentration was 10ppm.

If the final response was 75 then it would calculate that the the unknown concentration was 5ppm where 5ppm+10ppm=15ppm which is proportionate to 25response + 50response=75response.

Calculated as: (10ppm/(75response-25response))=RF=0.2 then RF*matrix Response =concentration in matrix; or 0.2*25=5ppm in the matrix.

That is assuming a linear response factor. If you add several incremental known concentrations(10ppm, 20ppm, 50ppm) and calculate response factors for each you can then plot the curve and see if the response is linear with concentration, if not then you have to solve for the curvature to find the unknown concentration which is a little more complicated.
The past is there to guide us into the future, not to dwell in.
Thank you for your replies.
19 posts Page 2 of 2

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry