We are using a Poroshell 120, EC-C18, 4.6 x 50 mm, 2.7 µm column on our new Agilent 1260 HPLC. Our mobile phase is 60%MeOH/40% Water. Run parameters are 1.5 ml/min flow rate with a 5 ul injection.
The backpressure after a newly inserted column is around 175-180 bar. This pressure consistently increases by about 3 bar every 100 injections or so. The peaks begin to broaden/tail once the pressure hits around 200 bar.
We were able to get about 3 months out of our first column (roughly 100 injections a day/5 days a week.) Once we noticed the peak shape starting to go bad, we replaced it with a new column. This one lasted about a month. We replaced that one with our current column and it has only lasted 1 week.
I am starting to believe there is another underlying issue here besides the column, yet when a new column is installed and calibrated, the peaks look great and pressure returns to normal (175-180 bar.)
Any other ideas what we may be looking at here?