I do not think that the answer to this is trivial, since you are possibly varying two things at the same time. On one hand, you are injecting the sample in a strong solvent (MeCN). At the same time, since it is dissolved in MeCN it is probably not in a salt form, but in a neutral form.
If a neutral non-ionizable analyte is injected in 100% MeCN, it will usually show peak fronting, if the injection volume is large enough. If an analyte is injected in a non-ionic form, and is converted by the mobile phase to the ionic form, you would usually get tailing, since the non-ionic form is more retained than the ionic form.
So: I think you got potentially two opposite phenomena going on in the proposed case.
hmm got a bit. Thanks sir. what if my analyte (glimepiride) is in 100 % methanol. Will it still be neutral (unionised)? i suspect it won't be neutral.
Will the pH of my analyte in diluent ( i checked it was 4.6) here effect the separation.
Mobile phase 20 mM Phosphate buffer (pH 3.0): ACN (40:60).
pKa of glime 6.2.
100 % methanol - less solvent strength than my worked mobile phase conditions. So ideally it should not effect as stated by Jiten.
Sorry to have so much confusion. I just want to know what role does pH of diluent play when we use buffer for mobile phase. ( both are different) How much it effects and what it does to peak.
Regards,
Amaryl.