My butanedione and pentanedione both tail badly on my column. I can run other stuff like alcohols,
ethers, aromatics and they look beautiful!
Running 6890/5973 MS and FID using split. Identical tailing at both detectors and at 10 and 1000 ppm.
Column: Agilent DB-624, 30M x 250um x 1.4 um thick. 43 cm/sec He. 1 uL injection (fast) in acetone.
Splitless injection; liner is deactivated 4 mm tapered with quartz wool.
40C for 1.5 minutes then 20 C/min to 200 C.
Inlet temp: Normally 250. Tried 200 and 160C. No effect on tailing.
Trimmed 2 feet on head of column. No effect on tailing.
Changed liner. No effect on tailing.
Cleaned liner assembly with gun brush and solvents. No effect on tailing.
Anyone have any thoughts on what else I could try?