weird sporadic detector noise

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

7 posts Page 1 of 1
On one of our purge and trap GC-MS systems we are seeing random noise peaks appear. By 'noise peak' I mean non-Gaussian TIC peaks made up of mass fragments of varying abundance (although all within 1000 counts of each other in any particular scan). They are non-centroidal so appear as e.g. 101.1, 101.8, 102.5, 102.7, etc. and there are peaks across the full scan range (e.g. between 30 and 250 m/z there are 220 peaks). The peaks affect a cluster of chromatograms in the sequence and appear approximately every 2 weeks. The peaks do not appear at any particular retention time, they are randomly dispersed throughout the TIC. I'll post a pic if I can find somewhere to host it.

This first happened after ion source maintenance. We assumed the EM or HED had been disturbed so we checked the connection and blew away any dust with an air spray. we didn't see the noise until a few weeks later and it has periodically appeared since then. I assume the detector is recording a signal intermittently within a run, but I don't know if it is an electronics issue, or if something is physically creating a signal.

Has anyone else had similar issues?

System is an Agilent 6890/5973 MSD.

Thanks
----suffers separation anxiety----
Hello

Run autotune...and see result. Perhaps dynode needs good cleaning.

Regards

Tomasz Kubowicz
Sometimes this will happen if you run the EM voltage too high.
I had that show up on a 6890/5973, although it wasn't intermittent. The high voltage lead (looks like the coil wire for a car) was arcing on one end. Trying to run a tune was futile.
The tune is fine. Can't see anything unusual, but the emvolts are up there - 2500 at the moment.
----suffers separation anxiety----
We see this when our EM is starting to die; sounds like a good time to replace it.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
This sounds very similar to a problem I am currently having and just posted about. Did you find the solution, was it the EM as mckrause suggested?
7 posts Page 1 of 1

Who is online

In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry