optimization of flow rate
Hi
I use a CHROMATOPAC for recording peaks in GC/FID.
how can I measure the W1/2 for calculating the effeciency?
I think that accuracy is very low.
thanks
happy new year
nina
Hello :D
Are you using CR8A? If so, what's the chart speed?
SK Srinivas, MPharm
CEO, K-Prime
Chromatography Training
Hi
I use C-R 6A chromatopac.
chart speed is 5.
happy new year
thanks
Hello:
If you're trying to calculate the plate count manually using CR6A, I'd suggest you take a printout of your chromatogram using a lower chart speed, say, 1 mm/min and use the lowest attentuation possible ( this will be the G key on the CR6A keyboard, the M key is for the chart speed)
You'll get large peaks in the chromatogram printout. This will make it easier for you to choose a peak and directly measure its W1/2 and calculate the plate count from that.
I hope this answers your question.
:D
SK Srinivas, MPharm
CEO, K-Prime
thank you very much.
best regards
In total there are 35 users online :: 0 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 35 guests