Hi
I use a CHROMATOPAC for recording peaks in GC/FID.
how can I measure the W1/2 for calculating the effeciency?
I think that accuracy is very low.
thanks
happy new year
nina

Hello :D

Are you using CR8A? If so, what's the chart speed?
SK Srinivas, MPharm
CEO, K-Prime
Chromatography Training
Hi
I use C-R 6A chromatopac.
chart speed is 5.
happy new year
thanks

Hello:

If you're trying to calculate the plate count manually using CR6A, I'd suggest you take a printout of your chromatogram using a lower chart speed, say, 1 mm/min and use the lowest attentuation possible ( this will be the G key on the CR6A keyboard, the M key is for the chart speed)

You'll get large peaks in the chromatogram printout. This will make it easier for you to choose a peak and directly measure its W1/2 and calculate the plate count from that.

I hope this answers your question.

:D
SK Srinivas, MPharm
CEO, K-Prime
Chromatography Training
thank you very much.
best regards
5 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry