Hi All,

While running a 1ul solvent (hexane) blank on a HP5890 series II I get strange jump in the baseline at about 4.5 minutes. The baseline jumps from 5pa to about 9pa and remians at 9pa throughout the duration of the run. Troubleshooting has pointed toward the injection technique, but this trouble started after running several samples using a consistent injection techniques and getting no strange jump. Anyone know what might be going on here or how to fix it? Thanks
andy

increase the split flow, minimise the column temp, change the glass wool, place right liner

vijay
india
Data system : event (such as set baseline) ?
Gas flow change at time (4.5 min): split flow or carrier saver changes?
Is vent flow high enough (greater than 25 or 50 ml )?

Sufficient contamination in injector may cause background jump if flows to / from injector changes during run.
Carrol
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry