Agilent 1260 RID Temp problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
hello,

get problem w/ the RID detector (Agilent G1362A), once the temp goes up by switching the method, it is hard to lower the temp down. ( it takes more than 3hrs to drop down from 35->30c ) do I suppose to turn off heater at this time (tried but no use) or keep the method setting of "Optical Unit temp is 30C", or it is just a HW issue for RID Temp controller? thank you!
That's a good question. I don't know if that RID has any cooling capacity; if it doesn't, and being well-insulated, just might take that long to drop temperature.

Our 1050 column heater is like that: can heat, but can only cool through heat loss.
Its normal. It doesn't have peltier element as JetStream column heater for example
Kind regards,
Ivan Styskin

LCCO Biotech
seems a HW issue, after changing the RID fan, it works (and can drop down to certain lower degrees!) but only hold for two weeks! doubt the board probelm, need to call service again!
4 posts Page 1 of 1

Who is online

In total there are 34 users online :: 1 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry