I am using DEAE sepharose FF Ion exchange resin as a first purification step in downstream processing (Large Scale). The column is packed with new resin, however now we are facing problem that column is not getting packed properly. We are getting very low number of plates (~500 for 16 cm Height). Also the bed height is continuously reducing at every packing. Can't figure out reasons for so low N value. As per slurry concentration (60%) we have not lost much of resin during repeated packing and unpacking of column. Now we are expecting as per new volume to be 15.5 cm but still wont get required hydraulic pressure at this height. We went further compressing the piston down to 14 cm to pack column at ~1.5 bar.
So my concern is
1) why we are getting so low N value, despite getting good As= 1.4.
2) why column height is continously reducing, does resin is getting degraded? Although resin is fairly new (only 20 cycles carried out).
3) What are reason for resin getting degraded? Does use of process water for long term storage have any effect?
4) What is oxidative degradation of ion exchange resin?