Gradient of Flow-Rate is Suitable?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Hi,

Is a gradient of flow rate suitable?

For example:

Time Flow Rate A% B%
0,00 1,00 10 90
5,00 1,50 10 90
10,0 1,50 50 50
15,0 1,00 10 90

Thanks!
in principal it is possible, but why you want to make your life that difficult?
Why you want to run a flow rate gradient???? If you want to get reproducibility - forget it.
Nice idea if you need a better baseline separation, but I would than use a different separation mode or minimum a different column.
Maybe some more details about your chromatographic conditions and let us know what you want to improve.
Gerhard Kratz, Kratz_Gerhard@web.de
I've only used flow gradient myself in the clean-out part of a gradient run, to increase throughput a little.

I'd say if it passes your system suitability, then OK. But have to wonder if you really need that.
As Gerhard mentioned, reproducibility will most probably be a nightmare. If you use UV-detection, remember that peak areas depend on flow-rate. Imagine what a flow-rate gradient means in this context...
If the flow gradient is only from faster column cleaning and initial mobile phase reconditioning than I think it's ok. If you want to use flow gradient to improve separation somehow it's not a good idea. There is so many variety in achiving good separation but the flow gradient it is not one of them :)
it won't do much anyway; since the flow-ramp is done at constant solvent strength, basically it's isocratic with the time-axis gradually squished as time increases, which could improve the looks of things because it would make the broader peaks at later retention times a bit narrower, but since it will also make them closer to one another, it won't actually improve the separation. There is an ideal flow-rate for the column, and if you can run at that flow rate without excessive back-pressure, then you should do so - for the entire method, and not just part of it. Modern columns, however, have pretty flat van Deemter plots (I believe), so if you can't, it doesn't matter (much). 50% changes in flow rate won't make much difference.

Yes, I've also used higher flows for the cleaning and/or re-equilibration portions of a run, usually motivated by a desire to shorten run time, and sometimes in the knowledge that I was using a slower flow not for the column, but for the detector.
6 posts Page 1 of 1

Who is online

In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry