In our lab, we are using a method from the US pharmacopeia. the mobile phase is a mixture of buffer:ACN (53:47,v/v). The buffer is prepared by dissolving 6,9 g of sodium dihydrogen phosphate monhydrate in 1L of water, then the pH is adjusted to 2.5 with phosphoric acid, after that, we dissolve 12 g of sodium sodecyl suflate in 1L of this solution.
This mobile phase is used for the assay of a basic drug.
When we mix the buffer solution with the ACN, small particules (probably of the ion pairing reagent) are formed.
We suspected that the concentration of sodium sodecyl suflate is too high to be dissolved in a mobile phase with a high organic proportion (47%ACN), so we tried to adjust this concentration with respect to the limits adressed in the pharmacopeia (-10%, 10,8g/l) but we obtained the same results as for the initial concentration.
I ask you kindly, if you have any idea about this problem, to answer the following questions:
- is the concentration of sodium sodecyl suflate too high to be dissolved in the mobile phase
- if this is true, why the concentration is so high in an USP method
- can we use a mobile phase with particules in suspension without harming the column or the HPLC instrument (we tried a few injections, but the substance was highly retained in the column and we got a peak with bad shape only when we increased the proportion of ACN to 57%, Rentention time was 21 min instead of 7min )
Thank you for your help