Things I've tried:
New septum. New liner. New ferrule on FID side of column (column nut was loose- can't figure out why, so I slapped on a new ferrule, no dice). FID is clean. Double-checked column "tail" length.
Tried different solvents- analyte in chloroform vs. analyte in methanol. Solvent peaks markedly reduced in both.
Attenuator set to 0.
Increased inlet pressure- no joy.
Changed out water trap on gas supply.
Tried different needles- no change.
Haven't changed the split ratio.
For a while, I was getting a normal-ish response maybe every 6th or 7th injection. Now I've gotten nothing but markedly smaller peaks for 3-4 days.
Could this be a problem with the jet? I've not torn down the FID that far before.