But, why do we get 25% higher results with the GC-FID? The comparisons are made twice, so I exlude that we have prepared our solutions improperly.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Because tetradecanol is a better internal standard than decafluorobiphenyl when the analytes are alcohols.We have two methods for determination of fatty alcohols (decanol, dodecanol, heptadecanol). The first one is for GC-MSD and the other based on GC-FID. The GCs are both Agilent 7890, and all chromatography settings are identical. The only thing that differs is the detection techniques and the internal standard. With MS system we use dekafluoro-biphenyl and with the FID (tetradecanol). The peak areas are used for quantifications. Specific mass fragments are used for quantification on the MSD. The peaks are pure, no interfering compounds.
But, why do we get 25% higher results with the GC-FID? The comparisons are made twice, so I exlude that we have prepared our solutions improperly.
We get cryptic questions (most of the necessary details omitted) - I give cryptic answers.Peter as usual is spot on but being a bit cryptic.