dblux_ » Sat Feb 28, 2015 9:34 pm
I have an Agilent 6890+5973N GCMS system that is interfaced with a Tekmar P&T system (autosampler).
The transfer line you want to remove comes from a concentrator.
To remove the P&T unit, is it as simple as taking out the transfer line and connecting the 2 swagelok fittings (1/16") together via a union to establish carrier gas to the injector?
Yes. Additionaly you will have to make changes in GC software, ie. change the sample source from P&T to ALS or manual.
It's not clear to me that there is any flow to the inlet of the GC that is supplied by the P&T unit. Anyone else out there ever done this?
The whole stream of gas in desorption stage comes from GC and gets back from the trap to GC inlet.