If I have a compound with a retention time of 4.5min, and I double the length of the column from 20cm to 40cm, does this mean that the retention time of the compound will be 9min?
-
- Posts: 715
- Joined: Wed Aug 04, 2010 2:45 pm
If your linear velocity is the same, nearly YES.
Gerhard Kratz, Kratz_Gerhard@web.de
-
- Posts: 58
- Joined: Fri Nov 21, 2014 8:31 am
with equation:
tr = L/ux
where,
tr - retention time (s)
L - length of the column (cm)
ux - velicity (cm/s)
If you double the length of the column, the retention time also doubles (if the velocity as Gerhard Kratz mentioneted is the same).
tr = L/ux
where,
tr - retention time (s)
L - length of the column (cm)
ux - velicity (cm/s)
If you double the length of the column, the retention time also doubles (if the velocity as Gerhard Kratz mentioneted is the same).
-
- Posts: 15
- Joined: Tue Dec 30, 2014 2:06 am
So does that mean that the resolution between two peaks will not change if the column length is increased?
-
- Posts: 715
- Joined: Wed Aug 04, 2010 2:45 pm
In theory - YES
Gerhard Kratz, Kratz_Gerhard@web.de
-
- Posts: 430
- Joined: Thu Nov 12, 2009 3:28 pm
Actually, if you are using isocratic elution, then doubling the column length should double the resolution. However, the additional time on column permits more diffusion to occur. Consequently peaks will be wider and so the increase in resolution will be something less than 100%. The peak maxima should be 2x farther apart, though.
-
- Posts: 161
- Joined: Tue Feb 10, 2009 4:15 pm
Yes, in theory the resolution will increase with sqrt(2)=1.4 (Since double column lenght, double number of plates)In theory - YES
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Only if you run isocratic.If I have a compound with a retention time of 4.5min, and I double the length of the column from 20cm to 40cm, does this mean that the retention time of the compound will be 9min?
Peter
Peter Apps
-
- Posts: 15
- Joined: Tue Dec 30, 2014 2:06 am
So if the resolution is between peaks is 1, and I double the column length, the resolution will improve a bit to 1.4?Yes, in theory the resolution will increase with sqrt(2)=1.4 (Since double column lenght, double number of plates)In theory - YES
-
- Posts: 161
- Joined: Tue Feb 10, 2009 4:15 pm
Yes (if isocratic)
-
- Posts: 154
- Joined: Thu Jan 20, 2005 11:27 pm
If increasing the resolution of two peaks is your main objective, I would try smaller particle size before using a longer column. A smaller particle size can sharpen the peaks, giving better resolution. Sometimes a shorter column with smaller particle size would be warranted. The column manufacturer can advise you on the column length and smaller particle size for your application once they know the original column and chromatography conditions.
11 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot] and 19 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Google [Bot] and 19 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science