I have to measure VPH with HS-GC-FID. I have problems with RSD (and SD).
I make 6 same concentration solution (from a normal alkan-mix, 75 ug/l for each compound) in headspace vials. I am using 20 ml vials with 10 ml water and 4g NaCl.
Headspace parameteres:
7694A HP headspace
1ml loop, 30 min equilibration time, 0,2 min loop pressure time, 0,2 min loop fill time, 0,8 min inject time, 70°oven temp and 85°loop temp, 20 PSI vial pressure, 12 ml/min carry gas from headspace
The carrier and vial pressure gas are N2
GC parameteres:
6890N Agilent GC
N2 carrier gas, 250°injector, 1:1 split, 1ml flow on column, oven: 35°for 1 min, then 12°/min to 140, 40°/min to 220.
I am using at the moment 25*0,2*0,5 column
I am getting for C5 to c7 5,2-6,2% RSD, and with C8 -C9 9-18% RSD.
It is like to be, i am getting discrimination, beacuse for c9, i get 42,8-61,1 aera, compare to c7 77,5-86,3 aera.
I tried to change liner, septum, o-ring, i changed the vial pressure (from 15 to 20PSI), loop fill time (0,1-0,2min) lowered the oven temp (on HS 80 to70°and on GC 40 to 35°), but it is still same.
Thanks for help in advance!
erdno