Then probably it's not detector gas flows related.
The simplest method of diagnosing would be to change the column. If new column has similar pattern then you would have to investigate further.
If new column has flat baseline then the old one is deteriorated.
BTW - this baseline is acceptable for me
So far, the samples' results are comparable with those before this happened. And the GC has to keep running. We can't pause for trouble shooting right now. But would like to try when things slow down, just out of curiosity.
Thanks all for help!