Iron contamination
Hi all,
I injected a sample that (apparently) had too much iron in it and contaminated my system. I've cleaned up the column and supressor now, but I'm wondering: how much iron can I have in a sample and still be safe to inject? My samples are about 20-30ppm iron, but I want to dilute as little as possible so I can still see the anions.
FYI:
Dionex ICS-2100
AS-18 column & guard
ASRS 300 supressor
CRD-200 carbonate removal
1mL/min 12mM-52mM KOH
Ideas?
Do you care at all about phosphate?
I assume that if you have that much of iron, you will not find any phosphate.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
H2Oh, I really don't care about phosphate.
I've also tried using the OnGuard II H cartridge but it doesn't seem to help, so I'm stuck with minimum 50x dilution.
Anyone?
SCX or else mixed bed guard?
In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)