I am trying to get a consistent detection for N2 and O2 using a Mol Sieve 5A column, He carrier gas, and TCD detection.
I am finding that when I inject 100 ul of room air, I get a nice separation of the two peaks, 7.8 min for O2 with area count of 500000 or so and 10.8 min for N2 with area count of 2e6, But when I try a second injection, the area counts drop to around 1e4 for O2 and 2e4 for N2.
I don't recall having this problem before and wondering if someone can suggest what might be the issue. If I monitor the total flow out of the split/splitless injector port, it is around 32 mls/min. If I monitor the flow out of the TCD outlet port, it is 18 mls/min with the ref and make-up gas. I don't think it is a flow issue, else why would I get such nice peaks with each first injection. The same thing happens if I start the oven at 40 oC or 50oC. And if I start the sequence, the first injection is great, all subsequent ones really low. If I restart the sequence, same thing happens.
Column retention issue? All suggestions would be most appreciated.