The analytes of interest will be in a solution of methanol, of which 75 uL is added with 150 uL of pyridine to 1 mL of BSTFA and allowed to react for 30 min @ 65C before injection. The issue I am experiencing is that 2 of our analytes, glycerin and dipropylene glycol, are co-eluting.
Column: TG-5MS 5% Phenyl, 0.32mm x 1um x 30m
Carrier gas: H2
Makeup gas: N2
Inlet temp: 220C
Split flow: 50 ml/min
Split ratio: 25
Constant flow at 2 ml/min
Oven: 40C- 250C at 25C/min, hold at 250C for 4 mins
Detector temp: 250C
Do I have any method options that will help with separating these components, or is my only option some change to the column? Thanks for any help in advance.