did you ever figure out how the frit caused the air bubbles? Did it cause degassing, or were parts of the assembly above the liquid level? The latter has caused problems several times in this lab. These so called last drop frits are often not tight, for instance one by VALCO has the 1/8" tubing to the pump just stuck into the teflon assembly holding the metal frit. Air has been drawn in between the tubing and the teflon assembly, among other problems.
Thus if I need to go to "the last drop" the intake frit is usually removed.
(Mentioned the manufacturer here as several complaints to a representative did not produce any response, so I wonder whether others noted such problems with what frits)