Ive got a problem with an Agilent GC. If the instrument is not used for a couple of hours and I run a 2 stage sample I get a result, but if I repeat the injection out of the same vial a further 6 times injection one is different to the other six. When the GC is calibrated in the morning it is therefore prone to failure if its not been used much during the night. Ive tried new septa, liner, gold seal, syringe and split line. If you do a hexane blank that doesn't cure it either. The instrument set up is as per the EP. Once its had one sample run on it it works fine but you cant justify ignoring a result.
Any help and advice would be much appreciated.