Benzalkonium Chloride Method
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
8 posts
Page 1 of 1
This is a pretty simple post, i hope. I'm looking for a Benzalkoium lc uv method that uses something other than a cyano or SCX column. Any help would be great.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Look for Chrompack application note #99, uses ion paring on a C18 column.
-
- Posts: 485
- Joined: Tue Aug 31, 2004 11:18 pm
We have a method for similar compound cetylpyridinium chloride with perfect symmetry and retention control. Check the following link, it also shows retention on C18 column. Our column is a mixed mode column with combination of reverse phase and ion-exchange groups.
http://www.sielc.com/compound_113.html
Compounds like benzalkonium tends to tail a lt due to residual silanol interaction, evan end-capped columns show tailing (in our comparisson it was leading brand C18 end capped column).
Contact us if you have questions.
regards,
Vlad
http://www.sielc.com/compound_113.html
Compounds like benzalkonium tends to tail a lt due to residual silanol interaction, evan end-capped columns show tailing (in our comparisson it was leading brand C18 end capped column).
Contact us if you have questions.
regards,
Vlad
-
- Posts: 208
- Joined: Wed Jul 06, 2005 3:35 pm
Hamilton Application Note #280. Uses a PRP-1 reversed phase column in straight RP mode.
Best regards,
Robert Haefele
rhaefele@hamiltoncompany.com
Best regards,
Robert Haefele
rhaefele@hamiltoncompany.com
-
- Posts: 485
- Joined: Tue Aug 31, 2004 11:18 pm
Dear Robert,
Can you provide a link to the method. I tried to find it on the website but did not see it.
you can email it to vlad.orlovsky@sielc.com
I want to compare it to our method (peak shape for quaternary amines)
regards,
Vlad
Can you provide a link to the method. I tried to find it on the website but did not see it.
you can email it to vlad.orlovsky@sielc.com
I want to compare it to our method (peak shape for quaternary amines)
regards,
Vlad
-
- Posts: 208
- Joined: Wed Jul 06, 2005 3:35 pm
http://www.hamiltoncomp.com/cgi-shl/pro ... applist280
Peak shape is actually not that great. Some improvement might be possible by running at elevated T and by using a 3 or 5um PRP-1 column. The application note is pretty old (>15yrs?) which explains the 10um particle size.
best regards,
Robert Haefele
rhaefele@hamiltoncompany.com
Peak shape is actually not that great. Some improvement might be possible by running at elevated T and by using a 3 or 5um PRP-1 column. The application note is pretty old (>15yrs?) which explains the 10um particle size.
best regards,
Robert Haefele
rhaefele@hamiltoncompany.com
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
You could also try the Acclaim Surfactant column. Featured on the homepage at www.dionex.com
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 267
- Joined: Tue Oct 12, 2004 9:35 pm
I have used amino phase for LC-MS of these compounds.
8 posts
Page 1 of 1
Who is online
In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science