I may have inserts somewhere. WIll check.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Now you know why when I did headspace I swapped out the loop for a 100 ul and ran with a direct connection to the columnHi Peter I just saw your last post. The aux EPC when setup with an Agilent 7694 has a bleed weldment of some sort just before the instrument. With the 7000 it goes to the manual flow controller which provides resitance and sets the flow rate as there is none in the aux epc they recommend a zero holed frit and then to a pressure regulator that is set to whatever the pressure you want the vial at. There is no point in having two pressure controllers in series, if you are gong to plumb through the Tekmar vial pressure controller anyway, then just take the gas straight off your house supply
SO you are saying that if I use the auxEPC it could overshoot the pressure when filling the vial and have nowhere to bleed to reach the pressure set point.That's about it - although I believe that some Agilent EPCs are just front pressure controllers, so if you have that kind you could bypass the Tekmar pressure control The flow controller is set to limit the flow rate through the needle so it builds up slowly perhaps so that won't happen but if it comes down to it I can always set the pressure on the aux epc higher and use the manual pressure valve to set the maximum allowed (setpoint) vial pressurization pressure same comment as in para 1 - one of the pressure controls will be redundant. If the EPC has a total flow setting (i.e. like a split splitless inlet EPC) you could use that to control the flow rate into the vial, and the pressure setting to limit the max vial pressure. I would be wary though that if the vial pressure overshot by even a little bit the back-pressure regulator would bleed some of the headspace out to control the pressure, leading to poor repeatability. There is another issue of where the EPC will read pressure - as far as I know Agilent puts its inlet pressure sensors in the septum purge line, which will not have a direct equivalent in the set up that you propose.
Man I had no idea what I was getting into with this instrument. I thought it would be simple cut the epc and hook it up.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.