Headspace Problem

Discussions about GC and other "gas phase" separation techniques.

24 posts Page 2 of 2
No the steam clean didnt make any difference...

I ran only two air and a clean water sample.
No standard

In every sample there was a 0,08pA Ethanol Peak.

Greetings
No the steam clean didnt make any difference...

I ran only two air and a clean water sample.
No standard

In every sample there was a 0,08pA Ethanol Peak.

Greetings
If you are still getting the same sized peak after a steam clean it points to persistent system contamination rather than just carryover from recent samples, and it looks likely that it is somewhere not affected by the steam clean, so the inlet is looking more and more suspect.

Peter
Peter Apps
I think it was the liner or the septa.....

I changed the liner to a broader one and the septa -> result was that the solvent peaks are smaller.

Carryover is now at 0,1 %

Gettin better
:)

Thanks guys
Good news !

Try increasing the septum purge flow. Also a double gooseneck liner might help even more.

Peter
Peter Apps
To increase the septum purge flow dont i have to run in splitless mode?

Double goose neck liner is ordered

i try a single goose neck that i found somwhere in my company
To increase the septum purge flow dont i have to run in splitless mode?
No, or at least not on any GC I have ever used.

Peter
Peter Apps
If you inject MeOH directly into the GC,and see only it, then you can rule out the injector as the caryover source.
I injected pure MeOH in the injector and there only was the MeOH Peak.....

....so the problem is the Headspace sampler.....

Thanks Bigbear
To increase the septum purge flow dont i have to run in splitless mode?
No, or at least not on any GC I have ever used.

Peter
The GC is a 6890, there is no option to increase the septum purge flow. The 7890 series allows the septum purge to be changed.

A 2 mm straight through liner without glass wool should be all you need (Agilent part# 5190-6168).
Thanks guys I´ll order the liner and test it
24 posts Page 2 of 2

Who is online

In total there are 149 users online :: 1 registered, 0 hidden and 148 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 148 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry