How reliable are the Waters Bourdon Transducers (600E and earlier)? I have a system that is spiking to high pressure on a water:methanol gradient. I really don't think the pressure is as high as the transducer is telling me. The same column did not exhibit as high a pressure on another system.

This is annoying since the system shuts down in the middle of the gradient.

It is tempting to just dial back the transducer to eliminate the shutdown, but I am hesitant.

Experience? Comments?

We had the same problem with 600 and ended up bypassing transducer

The different pressure on different systems amy also be due to differences in the tubing, rather than an incorrect reading of the pressure transducer.

The systems are nearly identical, and I try to keep it that way for comparison purposes.
4 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry