saccharose in HPLC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
8 posts
Page 1 of 1
i want to separate saccharose in H+ column. i tryed with H2O + 50 µL / L H2So4 and 55°C 0,6mL/min: 2 pics are coeluted! what is the ideal concentration of H2SO4? Thank you to answer me !
athebault
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
The cation exchange resins used for that type of separation are available in a variety of different ionic forms (H+, Ag++, Ca++, Pb++), each of which has a somewhat different selectivity. Decide which sugars you wish to separate and then contact the vendors (Bio Rad, Shodex, Dionex, etc.) for a recommendation.
Note that only the H+ form column is used with H2SO4 eluant; the others used only DI water (think about the solubility of Calcium sulfate!). That column genrally provides the least selectivity for sugars (it was originally designed to separate organic acids). The acid concentration has essentially no effect on sugar retention or selectivity.
Note that only the H+ form column is used with H2SO4 eluant; the others used only DI water (think about the solubility of Calcium sulfate!). That column genrally provides the least selectivity for sugars (it was originally designed to separate organic acids). The acid concentration has essentially no effect on sugar retention or selectivity.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 52
- Joined: Mon Jun 13, 2005 4:05 pm
Hi Annet,
What detector are you using? And what column? We do fermentation sugars here and use a Biorad Aminex 87-H column, works pretty good with an RI detector.
- Hubert
What detector are you using? And what column? We do fermentation sugars here and use a Biorad Aminex 87-H column, works pretty good with an RI detector.
- Hubert
-
- Posts: 357
- Joined: Wed Aug 03, 2005 1:15 pm
Anne,
Saccharose (sucrose) is EXTREMELY acid-labile; your chromatography is guaranteed to give you 100% conversion to a 1:1 mixture of glucose and fructose !!!
Suggest you do a lit. search; you need a non-acidic column and mobile phase, maybe a HILIC column with ammonium acetate or formate.
JMB
Saccharose (sucrose) is EXTREMELY acid-labile; your chromatography is guaranteed to give you 100% conversion to a 1:1 mixture of glucose and fructose !!!
Suggest you do a lit. search; you need a non-acidic column and mobile phase, maybe a HILIC column with ammonium acetate or formate.
JMB
-
- Posts: 357
- Joined: Wed Aug 03, 2005 1:15 pm
I should have said aq. ammonium acetate (formate) with MeCN for HILIC.
JMB
JMB
-
- Posts: 15
- Joined: Tue Aug 31, 2004 3:34 pm
JMB is correct about sucrose inverting under acidic conditions. If you are only analyzing for sucrose, you don't need the acidic eluent, and can use water to elute sucrose from the column. I've also seen a fairly sharp peak on ion-exclusion columns even with an acidic eluent if you test the column closer to room temperature, although you may still see some "shouldering". Most chemists that want to use the sulfonated gel type columns for analysis of sugars opt for the ligand exchange columns, usually in the calcium form. It is only if you need to separate weak organic acids in the presence of sugars that you would choose an ion-exclusion column.
-
- Posts: 7
- Joined: Mon Mar 07, 2005 2:05 pm
we use an RI detector and a biorad aminex 87-H columnHi Annet,
What detector are you using? And what column? We do fermentation sugars here and use a Biorad Aminex 87-H column, works pretty good with an RI detector.
- Hubert
athebault
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
As I indicated earlier, BioRad sells the HPX-87 columns in almost a half-dozen ionic forms. Each has somewhat different selectivity. The fact that the hydrogen form column separates the sugars that h.piatkowski is interested does not guarantee that it will separate the sugars you are interested in.
Check with BioRad to confirm that the H column is able to separate your sugars. If it is, then you have a dead or defective column. If it is not, then you have the wrong column for your application.
Check with BioRad to confirm that the H column is able to separate your sugars. If it is, then you have a dead or defective column. If it is not, then you have the wrong column for your application.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
8 posts
Page 1 of 1
Who is online
In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science