Hi Tom
I am responding to your comment on the leaking suppressor. Pete is assuming that you have a column switching valve interface.
The leakage issue is possibly related to
a) High backpressure from one of your lines. Measure the backpressure to the cell and ensure that this is less than 40 psi by trimming the backpressure coil.
Remember that all tubings post suppressor would contribute to the backpressure. For 4 mm applications we recommend that you use 0.01" black PEEK tubing for all connections upto the backpressure coil on the cell. From the cell backpressure coil to the suppressor and waste use 1/8" clear tubing.
b) Measure the system pressure A and then disconnect the line out of the eluent out port and measure the system pressure B.
The difference A-B should be < 100 psi. If this is > 100 psi and you have the correct backpressure then the high pressure may be from your cell.
c) If you are running external regenerant ensure that the regenerant flow is 3-5 ml/min. Use 1/8" tubing for making the connection to and from the suppressor. Reduce the applied pressure to the regenerant bottle if needed to ensure that the regenerant flow rate is within the recommended values.
d) Ensure that the suppressor is powered only with the eluent flow and regenerant flow on. If the suppressor is powered without flow you can damage the suppressor.
If you have any further questions feel free to contact me at:
kannan.srinivasan@dionex.com.
Cheers,
Kannan