TO-15 Analysis using the Marks UNITY2 and CIA Advantage

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Has anyone ran into trouble with the 1mL internal standard loop?

I have been trying to play with the split to get good recovery of my IS but when I do I overload the GC/MS. The best I have come up with is running a 10:1 split to get my IS around 500k counts. 5ppb runs around 5M counts with my 200ppb around 20M counts. I did go as far as 500ppb but the calibration cure started becoming quadratic.

I have talked with Marks and they can not swap out the IS loop for a larger loop (10mL would be nice)
The linear range is excellent at 80:1 split (5ppb to 1ppm) with good RSD and R squared values for most compounds but my IS is almost nonexistent.

Any suggestions?

Thanks,

Jones

Other info you might ask for:
Agilent 7890GC, 240 ion trap MS
Scotts Gas 1ppm IS.
Rxi-624 30m, .25mmID
Fixed the issue. I just got a 10ppm internal standard/surrogate. Duhh, K.I.S.S. :wink:

Jones
2 posts Page 1 of 1

Who is online

In total there are 126 users online :: 0 registered, 0 hidden and 126 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 126 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry