Hi,
I am using Hexane as a solvent and my analytes are a mixture of FAMEs.
Everything looks good when injecting in Split injection.
When I do split-less injection the solvent peaks covers my 1st analyte peak.
Remain the split mode reducing the split ratio untill you acquire a value that permits aceptable resolution (separation) between the solvent peak and first analyte peak. I wish you got nice elutions!!
You only want to go splitless if the sensitivity of your analysis is poor. Splitless should be your first attempt because transfer of the analytes to the column will be more controlled.
If you really need to use splitless in order to get the required peak sizes, then try a shorter splitless time.

Peter
Peter Apps
Thank you all.i did switch back to split.
If you can get the sensitivity you need, your results will be better and you'll be happier.
How far out is your first FAME and what is your oven initial temperature?

Best regards,

AICMM
7 posts Page 1 of 1

Who is online

In total there are 64 users online :: 3 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Bing [Bot] and 61 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry