Dear AICMM,
Thank you for your reply.
We decided to pull out the spring and push it in for another time because we noticed the FID was lit, but the button "SIG 1" on the right side of the machine showed value 0.0 after we made sure we lit the FID (we have a mirror to check the steam coming out of the flame). Also, the baseline did not response at the time we lit the FID (usually there would be a high peak at the time of igniting).
After pushing the spring harder to better connect with the FID collector, we tried to light the FID again. There is a number over 1.0 showed on the screen of the machine when we pressed the "SIG 1" button. Also, a peak showed up when we ignited the FID. So we think the signal was picked up by the spring and transferred to the peak on the real time window.
We checked and adjusted the air and hydrogen flow on the detector. They are 300 mL/ min and 30 mL/ min.
Then we decided to inject methane to check the carrier gas flow. But no peak came out within 15 min after trying a second time.
Any ideas why is there no peak for the methane?
You are not lit. Hydrogen and air flame should give you moisture out the castle of the FID so check it with a vial (or, ugh, I didn't say this, your glasses in your hand.) If not lit, use your ignitor even if not connected since it will choke off flows to help you light.
Set a high split vent flow using total flow. Your head pressure should read something if your system is plumbed correctly. Then set column flow (with H2, make-up and air off) somewhere close to where you want to be using the column head pressure knob to establish a back pressure on the column. Then turn everything else on and establish linear velocity. Then, when you calculate the column flow, turn up or down the total flow to establish your split ratio. Have to have enough split flow to be able to play with column back pressure.
You sound like you have a flow meter, you do right?
Best regards,
AICMM