- Posts: 2
- Joined: Thu Apr 14, 2011 8:01 pm
We recently obtained a used LECO 4D GCXGC-TOFMS. The system was last used in 2009 and was sealed when we obtained it. I get a very good vacuum of 1.2E X 10-7 torr. When I use source temp of 200C, it pass the leak test, it also pass the detector drift test. However, the system cannot pass the optimization with error message “mass calibration failed”. The tune calibration failed too. I replaced a new filament, it didn’t help. I tried to optimize the instrument many times in the past few months, it always failed at this stage. The detector voltage is around 1480V.
some of the error messages in the logfile stated "69.160 is out of tolerance, expected mass 68.995", 262.82 is out of tolerance, expected 263.99".
DAS test give message testing the high gain ADC, ADC bit test:expected 59 or 60 but read 61 at DAC 8500, expected 108 or 109 but read 110 at DAC6345. test low gain, the system freeze.
Appreciate any advice you can share.
Michael