we had a leaking problem from the pump assembly from the end of last year and that was when pressure dropping and pressure ripples started to happen. We managed to stop the leaking by replacing the fittings, and now the pressure seems to be stable. However, I noticed that the pressure never goes up to what it normally used to be before the leak(e.g. a particular gradient before the leaking is 1300 psi but now it is only 1100 psi).
At the moment I haven't observed any leaking from the pump assembly. The only cause on top of my head is that maybe the pump seal and the check valves are worn and torn. Does anyone agree with me? Could anyone suggest any other reasons for the pressure dropping? I don't want to blindly service the pump especially when I am not experienced at all, small problem can turn into big problem if you don't know what you are doing...
Another issue from the sample management system is- I noticed that whenever an injection is completed, there will be solvent built up on top of the HPLC vials, the carousel is very wet as a result of that. Is that an indication that the injector needs to be fixed?
I hope my questions are clear enough for people to understand and I do appreciate any suggestions for the problems I have. Thank you all very much for your time!