I'm having an issue with a gradient method. Setup as follows:
Kit:
-Waters e2695 Separations Module
-GE Mono-Q 5/50 GL Column
-Waters W2475 Fluorescence detector
Reagents:
-Mobile Phase A: Tris Base, CaCl2, PS-80
-Mobile Phase B: Tris Base, CaCl2, 1M NaCl, PS-80
Sample:
-Blood clotting factor (protein)
-Injection time 40mins
The assay starts at 75%A:25%B and ends at 30%A:70%B the goes to 100%B at the end of the injection for clean-up. It is this last step that it gets stuck at.
Usually I see 350~450psi but if I set the system to 1ml/min immediately after stalling it will hover between -14psi and 14psi. I then set the system to Line A, pressure goes up, back to Line B and we're back to normal.
It doesn't seem to coincide with a particular injection. It can occur after 3 injections or 43. very frustrating.
Any help would be much appreciated!!!