Inclined Baseline on Samples

Discussions about IC and related topics

2 posts Page 1 of 1
In our lab we are measuring trace chloride content of a particular compound and until recently we have been getting a incline in our baseline only when we run samples and in our blank. We experimented with different suppressor settings and nothing changed. Below are the parameters and any suggestions would be greatly appreciated.

Dionex DX-500 Ion Chromatograph
-Ionpac AS10Column, 4 X 250 mm (P/N 043118)
-Ultra ASRS Suppressor, 4mm (P/N 061561)
-Ionpac AG10 Guard Column, 4 X 50 mm (P/N 043119)
-Eluent: 100 mM Sodium Hydroxide
It appears that you are working under isocratic conditions. Is this correct? If so, can you elaborate on what you observe in terms of background? If you run two samples in series does the background resume where it left off on the previous sample?
2 posts Page 1 of 1

Who is online

In total there are 85 users online :: 0 registered, 0 hidden and 85 guests (based on users active over the past 5 minutes)
Most users ever online was 11068 on Sat Dec 06, 2025 9:55 pm

Users browsing this forum: No registered users and 85 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry