custom field
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
3 posts Page 1 of 1
How can i make a custon field for this formula:
Area Peak1 *100/Area Peak1 + Peak2
What you have written will reduce to area peak1 + area peak2.
Shall we assume you meant to ask for
area peak1*100/(area peak1 + area peak2)?
If only peaks 1 and 2 need be considered, name them, do not include any unknown peaks in your integration & calculations and include an area % field in your report. No custom field needed.
Thanks,
DR
the initial formula reduces to:
100 + Peak 2
and whatever formula you want to add the important question is: to which integration system?
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland