Baseline shift causing extraneous peaks - not reproducible!
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
10 posts
Page 1 of 1
What could be reason for a big baseline shift on only one sample in a sample set? I have one chromatogram with extraneous peaks and for the peak of interest the baseline also shifted below in the end valley. Several "extraneous peaks" were integrated, however, it appears to me that the extraneous peaks are not really peaks but the baseline shifting up and down. The peaks were not reproducbile. One note to be made is that the baseline shifted down slightly (from "0" to "-0.2") for the next chromatogramonly. After that, it went back to "0" and remained "0". Not a carryover from the previous injections either. The HPLC system was used everyday and the baseline was consistently stable for days after. What could be a reason for this? Please HELP!
-
- Posts: 39
- Joined: Thu Feb 14, 2013 8:54 pm
Could you provide a chromatogram so that we can see what these "peaks" look like? Also, a little bit more information about the method could be useful.
-
- Posts: 5
- Joined: Tue Nov 12, 2013 7:47 pm
Thank you for your attention. Could you please tell me how to include the chromatogram? I do not see any option to upload any images.
-
- Posts: 39
- Joined: Thu Feb 14, 2013 8:54 pm
-
- Posts: 5
- Joined: Tue Nov 12, 2013 7:47 pm
Here is the Chromatogram. Thank you!
-
- Posts: 39
- Joined: Thu Feb 14, 2013 8:54 pm
Alright great. Now that we have that, could you tell us more about the method. I.e. isocratic or gradient, mobile phases, etc.
-
- Posts: 5
- Joined: Tue Nov 12, 2013 7:47 pm
sorry...this is my first time using a discussion forum. I am also not so experienced in chromatography! It was an isocratic run with mobile phase 70:30 Phosphate Buffer:ACN / Wavelength 280 / Ambient Column Temp.
At this point, I am thinking it was an air bubble passing through causing the baseline shift.
At this point, I am thinking it was an air bubble passing through causing the baseline shift.
-
- Posts: 39
- Joined: Thu Feb 14, 2013 8:54 pm
That could be. Always check for air bubbles before running. A good indicator that there is an air bubble in the system is a drastic cyclic pressure change. What it also might be is carryover from a previous injection. One important way to tell the difference between an isocratic and gradient method is that isocratic method peaks will get wider as time goes on whereas gradient method peaks will have the same peak width, assuming that there is a constant change in %B per minute
-
- Posts: 5
- Joined: Tue Nov 12, 2013 7:47 pm
Thank you very much. That was a great help in moving towards the right direction.
-
- Posts: 2
- Joined: Thu Nov 21, 2013 12:38 pm
Maybe some debris washed off from the column frit.
10 posts
Page 1 of 1
Who is online
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 22 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science