I have run hplc for unknown plant compound using water:ACN (60%:40%), and I am getting lot of peak but when I am repeating it for its repeatability I am not getting the same RT and intensity.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Do you have any specific reason to use a preparative column for analytical purposes?Sir,
I have used C18 COLUMN and its dimention is 250* 20mm,10micro.
You may simply want to add a wash out step to the end of your run where you raise the organic composition of your mobile phase to 20%water:80%ACN. In many plant samples there will be oils and fats that will need more organic to elute them from the column. I analyze samples of broccoli seed extracts and we have to do this to remove the oil from the column or we get interfering peaks on following injections.hi,
I have run hplc for unknown plant compound using water:ACN (60%:40%), and I am getting lot of peak but when I am repeating it for its repeatability I am not getting the same RT and intensity.