C13 detecting issue on GC-MS Shimadzu QP 2010 Plus NCI

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

12 posts Page 1 of 1
Hi, I am setting the ISO 1948 method for dioxins and furans on GC-MS and I can not see the internal standard compound which have a C13 atom. I can easily separate and quantify the target compounds on GC-MS with negative chemical ionization, but the C13 I can not see any peak either on NCI, EI or pozitive chemical ionization.

Any thoughts?

Regards,
Irina
1. Are these "dioxins and furans" TCDD's and TCDF's ?

2. What are the actual RT values for the analytes and the expected RT for the 13C I.S. ?
1. Are these "dioxins and furans" TCDD's and TCDF's ?

2. What are the actual RT values for the analytes and the expected RT for the 13C I.S. ?
1. yes they are

2. from RT 17 to 29 for analytes and I was expecting appropriate RT for 13C compounds. But even if I inject only 13C labelled compound and look for both scan and their ions I can not see any peak and I do not know why. I have the same column as in the ISO (TG-5MS TraceGold 30m, 0.25 mm, 0.25 um), injection temperature I tried it from 100 to 330 and still no peak. I'm stuck and I can not waste to much of the solution because they where like 200 ul for each calibration point and the cost was about 3000 euro.
I would suspect that your 13C internal standards are possibly bad; do you have a manufacturer certificate of analysis with GC/MS chromatogram and spectrum for each IS ??
Double check the concentrations of your labelled standards, and the dilution (is any) that you used to make your working standards.

Peter
Peter Apps
The standard staid a few months opened, but on their certificate expiration date is on 2017. I only have a chromatogram with both labelled and unlabeled standard where it said they are showing up at the same RT as the unlabeled.
I did not use any dilution because it was a set of 5 calibration point, an extraction solution (to add when start extracting the samples) and a sampling one which included only the IS at a concentration of 0.1-0.2 microg/ml.
Irina
Since,

1) your GC/MS system is operating correctly,
2) you do not perform any dilution of the IS solution,
3) but the I.S. is below LOD

I think that we have to assume that the I.S. solution is below the concentration specified by the manufacturer.

I know the I.S. is very expensive, but to resolve the issue you might consider injecting an aliquot that has been concentrated by a factor of at least 10; whatever the result, you would then have some solid evidence to provide to the mfr.

By the way, are using full-scan MS or selected ion monitoring in your method ? SIM would certainly give better sensitivity than full-scan data acquisition.

One last caution, before you challenge the mfr. check ALL of your exptl. details, then re-check ALL of your exptl. details, then have a third party check ALL of your exptl. details.
One more thought, test your GC/MS system to see that it meets the Shimadzu GC/MS sensitivity specification before you consume any additional 13C standards.

If not, apply the usual techniques---have a clean glass liner/new septum/ cut 1-2 cm from head of column/clean the ion source/clean the analyzer etc.

Good Luck !!

JMB
Do you have standards of unlabelled analytes where you can clearly see good peaks ? What is the concentration of those standards and how does it compare to the concentration of the labelled standards. ?

Peter
Peter Apps
Peter,
I have good chromatogram for unlabelled compounds and the concentration is from 0.005 ug/ml to 0.2 ug/ml.
JMB,
I cleaned up again the ion source, change the liner and septa and now I am waiting for the vacuum to be stable and I will check again the IS and I will concentrate it also.
Thank you all for reply. I will let you know how it works.

Irina
The concentrations are very low - the maximum that you are putting on column (assuming a 1 ul splitless injection) is 0.2 ng.

Have you directly compared the response for labelled and unlabelled compounds at the same concentration ?

Peter
Peter Apps
I think the column was the problem because after I changed it with a new one I can see the peaks for C13 also. I know the concentrations are very low and that's why I am analyzing only in SIM mode and I didn't see the bleeding of the column.

Thank you all for your suggestion
12 posts Page 1 of 1

Who is online

In total there are 55 users online :: 0 registered, 0 hidden and 55 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 55 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry