Hey ,
I am Lopa . I am using HPLC in gradient mode.My mobile phase solvents are water+TFA(0.2%) and methanol+TFA(0.2%).
But TFA makes the two solvents pH 1.56 and 0.67 respectively.
My column is a c18 column which has pH range up to 2-8 .Can i use this mobile phase?

Please suggest me
Hello, Lopa,

Most C-18 materials have the C-18 alkane chain attached covalently to a silica surface via a silane: Si-(CH2)17CH3. At pH 2.0 or less, the Si-C bond will start to hydrolyze. Consequently, sustained use of 0.1% TFA results in a steady loss of capacity of a C-18 column over a period of several months. With your 0.2% TFA the rate of loss will be even faster.
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
In Addition to Andy's comment please check if your column is a so called trifunctional or polymeric bonded packing material, or a monofunctional bonded Phase. Under acidic conditions trifunctional bonded phases in General have longer lifetime. But definitively C18 chains are stripped off faster with 0,2%TFA. What is the pK of your target compound?
Gerhard Kratz, Kratz_Gerhard@web.de
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry